Protocol:The Omalizumab ELISA kit is designed to measure free Omalizumab with high specificity and sensitivity . This assay employs the sandwich enzyme immunoassay technique. A precoated anti-Omalizumab 96 well plate is provided. Calibrator, quality control samples and test samples are pipetted into the appropriate wells. Omalizumab present in biological matrices is bound by the immobilized capture antibody. After washing away any unbound substances, enzyme linked detection antibody is added to the wells. The plate is washed to remove any unbound antibody-enzyme reagent and a substrate solution is added to the wells for color development. The color development is proportional to the amount of Omalizumab present in test samples and the concentration is calculated from the standard series.
Methode type:Direct sandwich ELISA
detectionMethod:Peroxidase / OD450
Principle:
detectionMethod:Peroxidase / OD450
Assay time:2.5 hours
Specificity:Omalizumab
Assay procedure:This assay employs the sandwich enzyme immunoassay technique. Anti- Omalizumab is coated onto a 96 well microplate. Calibrator and test samples are pipetted into the appropriate wells. Omalizumab present in biological matrices is bound by the immobilized anti- Omalizumab antibody. After washing away any unbound substances, enzyme linked anti- Omalizumab antibody is added to the wells. This antibody is developed and purified specifically against Xolair® (domain residing in Fc portion of the Xolair® molecule). The plate is washed to remove any unbound antibody-enzyme reagent and a substrate solution is added to the wells for color development. The color development is proportional to the amount of Omalizumab present in test samples. The color development is stopped and the intensity of the color is measured
Reagent preparation:Prepare only the appropriate amount of required reagent on the day of use. Store all reagents as per instructions stated on the label. 1. Wash Buffer (1X) Preparation: Dilute wash buffer concentrate with ultra-pure water 1/10 before use (for example add 20mL concentrate to 180mL ultra-pure water). Mix well. 2. Detection Reagent (1X) Preparation: Dilute detection reagent with assay buffer 1/1000 before use (for example add 11μl concentrate to 11ml of assay buffer). Mix well. 3. Preparation of Calibrators: Prepare calibrators with concentrations ranging from 2,500 ng/mL to 78 ng/mL. The following is an example calibrator curve.
Results calculation:1. Construct a standard curve by plotting the absorbance obtained from each standard against concentration. Use a 4 or 5 parameter curve fit. Alternatively a log-log curve fit may be used. 2. The concentration of the unknowns can be read directly from this standard curve using the absorbance value for each sample. 3. Any sample undiluted or diluted still reading greater than the highest standard should be diluted appropriately with calibrator diluent and retested. If the samples have been diluted, the concentration determined from the standard curve must be multiplied by the dilution factor